Design and Simulation Study of a CNT-based Multisource Cubical CT System for Dynamic objects
时间:2018-09-19 16:35 来源:未知 作者:admin 点击: 次
Abstract:The purpose of this paper is to design and simulate a new computed tomography (CT) system with a high temporal resolution for dynamic objects. We propose a multisource cubical CT (MCCT) system with X-ray tubes and detectors installed on a cube. Carbon nanotube- (CNT-) based X-ray focal spots are distributed on the twelve edges of the cube. The distribution of X-ray focal spots and detectors completely avoids mechanical movements to scan an object under inspection. CNTs are excellent electron field emitters because the use of a “cold” cathode makes it possible to fabricate a cathode with multiple electron emission points, and the CNT-based X-ray focal spots possess little response time and programmable emission. The proposed rotation-free MCCT system can acquire a high scanning speed when using a high frame rate detector. A three-dimensional (3D) reconstruction algorithm with tensor framelet-based L0-norm (TF-L0) minimization is developed for the simulation study of the MCCT. Simulation experiment results demonstrate the feasibility of the MCCT system.
1. Introduction
Computed tomography (CT) is an important nondestructive testing tool which has been widely applied to medicine, industry, biology, and so on [1–5]. In the field of medicine and biology, dynamic objects such as small animals and human hearts are the subjects most often investigated. It is known that cardiovascular disease is a serious menace to human health. On this account, small-animal imaging plays an important role in providing a basic understanding of the mechanism of the disease, drug discovery, clinical assessment, and so on [1]. However, physiological motions of small animals such as a mouse are much faster than those of humans, and motion-induced artifacts inevitably degrades the CT images [6]. CT and other imaging modalities have revolutionized clinical diagnosis. However, dynamic CT is practically difficult because of the requirements for high spatial resolution and temporal resolution. The challenge in scanning a dynamic object is the lack of real-time detection technology to monitor the rapid changes and alleviate the motion-induced artifacts.
To abate the motion-induced artifacts, a natural idea is to accelerate the scanning process. Since CT was introduced to clinical medicine, some researchers focused on increasing the temporal resolution using traditional CT systems. Previous studies suggested that the temporal resolution should not exceed 50 msec for cardiac CT [7]. For a single-source CT system, the source and the detector installed on a gantry simultaneously rotate around the object under inspection to acquire complete projections. The limitation in rotation speed makes it difficult to further increase the temporal resolution, though slip-ring technology greatly reduces the scanning time dominated by interscan delays [8].
Multisource CT can be used to improve the temporal resolution. In 1979, the first multiple-source CT system for dynamic objects was designed at the Mayo Clinic [9], consisting of a large number of X-ray sources and fluorescent screens. This idea has a profound impact on CT. Liu et al. [10] proposed a three-source and five-source micro-CT system and applied it for a half-scan Feldkamp-type reconstruction, which can deliver a 2.33 and 3.18 speedup factor, respectively. Kachelrieß et al. [11] suggested a spiral CT with X-ray sources and detectors for cardiac imaging. An approximate cone-beam algorithm was used for CT reconstruction. The temporal resolution is roughly a factor of better than that with a single-source CT scanner. In 2005, Siemens designed a dual-source CT system which produces high temporal resolution images without the loss of spatial resolution and halves the scanning time [12, 13]. Liu et al. [14] proposed a multisource X-ray interior imaging system to achieve an ultrafast scan for small animal imaging. This system indicated that a 50 msec scan speed is possible without gating at low dose. Wu et al. [15] proposed a swinging multisource CT system for aperiodic dynamic imaging in industrial applications.